Technical Documentation 3 answers

How to comply with EU 207/2012 electronic IFU requirements for medical device apps?

Anonymous · Published February 18, 2026 · 1 comment
How should manufacturers of standalone software medical devices (apps) distributed via app stores comply with the requirements of EU regulation 207/2012 regarding electronic Instructions for Use (eIFUs), especially the requirement to provide paper versions?
What are the expectations of Notified Bodies in such cases? Are there any practical solutions or experiences from others about providing paper IFUs for software-only apps?

Join the discussion. Leave a comment. Guest comments are welcome — add your email to get reply notifications.

Anonymous 4 months ago
This is a common issue for software medical devices. Looking for practical solutions.
Reply to this comment

Discussion

3 Answers

Accepted answer Dr. Oliver Eidel · Founder & CEO, OpenRegulatory ·
One possible solution is to embed the IFU in the app and/or provide it on your company website, with a link in the app itself. Using the eIFU symbol in the app can help. In my experience, Notified Bodies often accept this approach, especially when it's clear how users can access the IFU.

Join the discussion. Leave a comment. Guest comments are welcome — add your email to get reply notifications.

No comments yet. Be the first to share your thoughts.

A
Anonymous ·
My suggestion is to state in your internal procedure that the paper IFU will be delivered upon request. In practice, this request rarely happens, so it may only be a theoretical issue.

Join the discussion. Leave a comment. Guest comments are welcome — add your email to get reply notifications.

No comments yet. Be the first to share your thoughts.

A
Anonymous ·
From what I've seen (having worked as an NB auditor), you should still address the requirements of 207/2012, even if they don't make much sense for some products. Typically, companies prepare a matrix documenting the rationale for each requirement's applicability, with supporting evidence from risk management and QMS procedures. It's a regulatory requirement, so depending on your situation, you could get a minor or major nonconformity if not addressed. There's a draft of an updated eIFU regulation, but it's not published yet.

Join the discussion. Leave a comment. Guest comments are welcome — add your email to get reply notifications.

No comments yet. Be the first to share your thoughts.

Want to add your answer to this question?
Write an answer under your name by logging in or signing up, or post anonymously.

Still have a question? Ask a question here publicly — for free.

Or talk to one of our consultants — first calls are free. Check out our services and prices.

Looking to automate your regulatory work? Check out our eQMS, Formwork. Built for lean, founder-led companies. There’s a free version too.