Quality Management 6 answers

Is a new QMS audit required for MDR certification if we already have MDD and ISO 13485 certification?

Anonymous · Published February 21, 2026 · 1 comment
Our company has a device certified under the MDD with a certified QMS (quality management system). We are planning to transition to MDR certification.
Do we need to undergo a new QMS audit according to MDR, or is it sufficient to update and submit our technical documentation to our notified body? Specifically, does an existing MDD or ISO 13485 QMS audit meet the MDR requirements, or is a new audit mandatory?

Join the discussion. Leave a comment. Guest comments are welcome — add your email to get reply notifications.

Anonymous 4 months ago
Can you specify your device classification? The requirements may differ for Class I versus higher classes.
Reply to this comment

Discussion

6 Answers

Accepted answer Dr. Oliver Eidel · Founder & CEO, OpenRegulatory ·
For our company, the stage I audit was a verification of our QMS specifically against MDR requirements, and then we sent in our technical documentation for review and CE marking. So, yes, you will likely receive a QMS audit because the QMS requirements under MDR are more specific and stricter than under MDD, like reporting timelines. Basically, a new audit is expected.

Join the discussion. Leave a comment. Guest comments are welcome — add your email to get reply notifications.

No comments yet. Be the first to share your thoughts.

A
Anonymous ·
What conformity assessment do you have? If you have at least class IIa you also need to have a certified QMS.

Join the discussion. Leave a comment. Guest comments are welcome — add your email to get reply notifications.

No comments yet. Be the first to share your thoughts.

A
Anonymous ·
We have a certified QMS, and I asked a similar question. It seems best to check with your notified body. In our case, the notified body said the MDR audit is more strict for the QMS, so if we decide to market an MDR compliant device, we need a longer audit than just a surveillance audit.

Join the discussion. Leave a comment. Guest comments are welcome — add your email to get reply notifications.

No comments yet. Be the first to share your thoughts.

A
Anonymous ·
I discussed this with our notified body. Neither a MDD nor a ISO 13485 QMS audit can formally be used for MDR QMS. The conformity assessment procedure for MDR that needs a certified QMS must be based on its own audit. For example, our company was quoted three days for ISO 13485 audit, but five days for MDR QMS, so MDR is more detailed. It’s always good to double check with your notified body though.

Join the discussion. Leave a comment. Guest comments are welcome — add your email to get reply notifications.

No comments yet. Be the first to share your thoughts.

A
Anonymous ·
Some notified bodies use the MDR as a reason for a new certification, but our notified body didn’t mention anything about that. We wanted to change notified bodies, but the new one did not accept the QMS certificate, so it would need a new QMS certification. Since ISO 13485 is harmonized, it should be okay with MDR. QMS is audited annually and, together with the technical documentation, there will be an on-site audit, but I don’t think an NB will make a complete new first certification except for the three-year re-certification.

Join the discussion. Leave a comment. Guest comments are welcome — add your email to get reply notifications.

No comments yet. Be the first to share your thoughts.

A
Anonymous ·
From our experience, some notified bodies treat the transition to MDR as a requirement for a new QMS audit, while others may be more flexible, especially if your ISO 13485 is up to date. It’s best to check directly with your notified body, as practices seem to differ.

Join the discussion. Leave a comment. Guest comments are welcome — add your email to get reply notifications.

No comments yet. Be the first to share your thoughts.

Want to add your answer to this question?
Write an answer under your name by logging in or signing up, or post anonymously.

Keep reading

All questions
A

Is it necessary to have technical documentation ready before certifying the QMS under MDR for a class IIa device?

Anonymous · 2 comments
12 months ago
We are preparing to certify our Quality Management System (QMS) but do not yet have all of the technical documentation for our medical device completed. According to MDR 2017/745 (Annex IX, Chapter 1,
Accepted 4 answers

Certification to ISO 13485 is independent from being a manufacturer under MDR Annex IX. Many certification bodies are not Notified Bodies for MDR, so

A

Is it necessary to have technical documentation ready before certifying a QMS under MDR for a class IIa medical device?

Anonymous · 1 comment
12 months ago
Our company wants to certify its Quality Management System (QMS) as soon as all documentation is ready, but we do not yet have all the technical documentation for our medical device completed. Based o
Accepted 1 answer

It's generally fine to proceed with certifying your QMS before the technical documentation for your device is complete, especially if you are aiming f

A

How to generate EUDAMED ID DI for MDD devices and what manufacturer device code is needed?

Anonymous · 1 comment
4 months ago
When registering MDD devices in EUDAMED using the EUDAMED ID DI, is a manufacturer device code required, and where can it be obtained? Does EUDAMED generate the EUDAMED ID and DI automatically, or doe
Accepted 7 answers

After ISO 13485 certification, you should close out any audit findings and address them in your Corrective and Preventive Actions (CAPA) process. Expe